    Thu, 29 Sep 2022 23:14:15 GMT – Probes for MDO3014

    I'm happy to help you find probes for your MDO3014.You need probes for a spectrum analyzer, and I'm assuming that you are referring to the built-in Spectrum analyzer on the MDO3014 unit. If so, Tektronix actually has an FAQ that covers this question, linked below.
    You also may be interested in the TPA-N-VPI Adapter. This adapter accepts probes with BNC, TekProbe™, and TekVPI™ connection interface. If you are currently using a probe with one of these interfaces, this adapter would allow the probe to be used with N connector of the MDO3000.

    Links:
    What probes are supported on the RF input of the MDO3000 and MDO4000 series?: https://www.tek.com/en/support/faqs/what-probes-are-supported-rf-input-mdo3000-and-mdo4000-series
    TPA-N-VPI Probe Adapter: https://www.tek.com/en/manual/tpa-n-vpi

  • RE: Looking for DAQ6510

    Thu, 29 Sep 2022 23:14:14 GMT – Looking for DAQ6510

    We suggest the DAQ6510 with 7706 module. You'll need the KTTI-TSP option and custom script.

  • RE: Upgrading AWG5208 sample rate

    Thu, 29 Sep 2022 23:14:12 GMT – Upgrading AWG5208 sample rate

    Option AWG5200-8-2550 is the upgrade from 2.5GS/s to 5GS/s.

  • RE: P2220 Probe and MDO34

    Thu, 29 Sep 2022 23:14:11 GMT – P2220 Probe and MDO34

    It will work. Just a manual set up to tell the scope that a 1x probe is connected and they will be good to go.

  • RE: e*Scope Password for MSO44

    Thu, 29 Sep 2022 23:14:09 GMT – e*Scope Password for MSO44

    The MSO44 should not have any e*Scope password by default. If you need to clear the password, you can do so by going to Utilities -> IO and selecting the LAN Reset button. This will set all of your network preferences to the defaults as well, so you will likely need physical access to the scope to setup the network again.

  • RE: TPA-BNC Adapter and MSO54

    Thu, 29 Sep 2022 23:14:08 GMT – TPA-BNC Adapter and MSO54

    Yes, the TPA-BNC adapter is compatible with the MSO54.

  • RE: Question about AWG5200-8-2550

    Thu, 29 Sep 2022 23:14:06 GMT – Question about AWG5200-8-2550

    AWG5200-2-2550 is an upgrade option on a 2 channel AWG5202. It increases sampling rate from 2.5 GS/s to 5 GS/s (10 GS/s interpolated)

  • RE: Time Base Increments of TBS2074B

    Thu, 29 Sep 2022 23:14:04 GMT – Time Base Increments of TBS2074B

    I am contacting you in regards to your request with Tektronix about time base increments of the TBS2000B. The time base specifitations can be found on page 12 of the performance verification manual. I have linked this below:
    • TBS2000B Performance Verification Manual: https://www.tek.com/en/manual/oscilloscope/tbs2000b-series-oscilloscopes-tbs2000-basic-oscilloscope

    The time base increments are in a 1-2-4 sequence from 1ns/div to 100sec/div.

  • RE: Volt/Div settings for MSO44

    Thu, 29 Sep 2022 23:14:03 GMT – Volt/Div settings for MSO44

    Here are input sensitivity ranges:
    1 MΩ; 500 µV/div to 10 V/div in a 1-2-5 sequence
    50 Ω: 500 µV/div to 1 V/div in a 1-2-5 sequence.

    Note: 500 μV/div is a 2X digital zoom of 1 mV/div or a 4x digital zoom of 2 mV/div, depending on the instrument bandwidth configuration.

    Also a regarding a 1-2-5 sequence, following patterns helps explain what a a 1-2-5 sequence is: 1mV,2mV,5mv,10mV,20mV,50mV, 100mV, 200mV, 500mV, 1V, 2V, 5V, 10V /div.

  • RE: Calibration for the TPA-BNC?

    Thu, 29 Sep 2022 23:14:01 GMT – Calibration for the TPA-BNC?

    There is no Calibration for the TPA-BNC; there is only a Functional check. If the probe your using it with works, talks to the scope, and meets specifications, the TPA-BNC is working.
    Our service department does offer this Functional check if needed, at a cost of about $265.

    To set up a repair RMA (model and serial number)
    call 1-800-833-9200 opt 2. Or 1- 800-438-8165
    Or send an email to rmarequest@tektronix.com
    Or https://www.tek.com/en/service-quote
    Or https://www.tek.com/en/services/repair/mytek_rma_request

  • ‹
  • 1…
  • 39
  • 40
  • 41
  • 42
  • 43
  • 44
  • 45
  • 46
  • 47
  • …58
  • ›
This site is for Tektronix Customers to Learn, Share and Collaborate.
 

© 2025 TEKTRONIX, INC. | Terms of Use | Code of Conduct | Contact us

▼