  • RE: diagnostic error 313

    Tue, 04 Oct 2022 23:15:50 GMT – diagnostic error 313
    Error 313 could be a Demux on the Acquisition board. If the error continues to count up every time the diagnostics is ran, then a part or parts has failed, if only shows sometimes a part or parts maybe failing. This will need to come into our Tektronix factory service center.
  • diagnostic error 313

    Tue, 04 Oct 2022 23:15:00 GMT – diagnostic error 313
    DPO70K scope is showing diagnostic test error acquisition 313. What is this indicating.
  • RE: Scope and TDP0500 Firmware

    Tue, 04 Oct 2022 22:32:28 GMT – Scope and TDP0500 Firmware
    We need to know the scope model in question, and verison of code on it now.
  • Scope and TDP0500 Firmware

    Tue, 04 Oct 2022 22:31:13 GMT – Scope and TDP0500 Firmware
    Need to see if our instrument needs a firmware update to support the new TDP0500 differential probe we purchased?