RE: Transfer Recall Settings to DPO 2024B
The TDS2024C and DPO2024B do not have compatible .SET file formatting, nor do we have a software update or program to do so.<br> <br> You could open the TDS2024C .SET file in VIM and manually compare it to the DPO2024B .SET file, then re-fill the fields that are applicable, but they're quite different in terms of formatting. -
RE: How to get screen caputure from DPO7000 series using VBA
Try setting the timeout to be long enough to allow the full image data to be sent. Depending on how you're connected (in this case GPIB) you may need a longer timeout.<br> <br> osc.timeout = 3000 -
RE: MDO 4000 Series
To add the measurements, select Measure->Add Measurement-->(Select Source)-->Measurement Type.
From there, you can add a frequency measurement or scroll with the multipurpose knobs to see other options like positive and negative pulse width.
These measurements use the portion of the acquisition selected in the gating menu.
You may change the region you're measuring in with (Under Wave Inspector in the front panel) Measure --> More --> Gating.
-
RE: Which command group to retrieve the specturm measurement results of MSO44?
Hi Jeremy,
The image you attached is also the algorithm used by the scope (page 635 in the MSO4 help manual).
Looking at the equation, the values that you need to calculate channel power would be CBW (channel bandwidth), RBW (resolution bandwidth), N (number of data points) and Pk (power at the kth data point.)- To set/query CBW (the difference between the upper and lower frequency)
CHx:BANdwidth
Correction:MEASUrement:MEASx:CPWIDTh
- To set/query RBW (must set RBW mode to manual first)
SV:RBWMode MANual
SV:RBW - To find N and Pk, you could save the spectrum data and then parse through each point. The Save and Recall command group (2-92 in Programmer Manual) should have the SCPI commands to set up where to save and such. You can also change the record length;
HORizontal:MODE:RECOrdlength
You can look more at the acquisition or data command group for other parameters you might want to change.
Hopefully this will be a good starting point. I will let you know if I have any more ideas for working around the bug or edit if I find some more suitable commands for finding this programmatically.
-Iris - To set/query CBW (the difference between the upper and lower frequency)
-
RE: MSO5 series spurious tone
The other two things are also normal in terms of digital signal processing and how spectrum analyzers work.
Along with this, changing the horizontal/vertical scale on the MSO5 series moves both the waveform and graticule relative to the display position, which may be something you're not used to on other scopes. -
RE: MSO44 scope SCPI trigger questions
Hi,
You can use the A and B triggers to set up triggering on sequential events for example, where the A trigger acts as a qualifier. You can read more about their dual usage in page 134 of the MSO 4 series user manual.
As for the SCPI commands you are looking for;- Setting trigger source to channel 3 (example using edge trigger type):
- Setting trigger level to 50% can be done with;
Furthermore, if you want to know what the trigger parameters are after setting it to 50%, you can also query using (example using trigger A):
TRIGger:A?
-Iris
-
RE: Catalogues request for Oscilloscopes
Hi Nathalia,
Please see the Tektronix Sales Contact page in order to get in touch! We will take a further look into your requirements and offer services such as demos and technical support from here.
Thank you for reaching out. -
RE: MSO44 scope SCPI questions
Hello,
- To measure from other channels, you should first select it using this command below(for example) before running the MEAS:addmeas command.
DISPLAY:SELect:SOUrce CH2
DISPLAY:SELect:WAVEViewx:SOUrce CH2
- To check the "show stats badge" and change the limit to 100, please use these commands replacing x with whichever measurement number you want to adjust the population limit for.
MEASUrement:MEASx:DISPlaystat:ENABle ON
MEASUrement:MEASx:POPUlation:LIMIT:STATE ON
MEASUrement:MEASx:POPUlation:LIMIT:VALue 100
Hopefully this helps.