RE: DMM6500 pops up an error code 1135 when using remote control
Most likely there is a syntax error in the SCPI command you are sending.
On these touch screen models like DMM6500, you can enable command logging in Event Log and have the instrument identify the offending command for you.
Here is brief video showing the process:
https://www.youtube.com/watch?v=cVU_C308EUc -
RE: Reading measurement data from Keithley SourceMeter 2602A
Hello,
Also, the time allocated for the measurement, needs to be less than the Pulse Width.
As a test, set the smuX.measure.nplc = 0.001
This is the minimum allowed value for 2600X.
When you say you do not get all values:
Is the sweep completing? Or is the sweep "stalling" because maybe there is a timing problem with nplc, pulse width and pulse period.
The print(smuX.nvbufferX.n) will indicate how many values are in the buffer.
Let us knw if expected number of readings are in the buffer, but you are not successfully transferring all the data.
I take it you using LAN connection now? -
RE: 主线程卡死
Hello,
Seems you are attempting to use NI VISA from multiple threads in C#.
Can you clarify which exact model of instrument you have?
Also, can you give us an idea of the operation you are attempting with that instrument.
Sometimes a multi-point sweep can take a long time to run and in that case it would be expected that attempts to read the information would not be processed until after the sweep operation completes.
This NI forum post seems to have some good info about using VISA from threads:
https://forums.ni.com/t5/Instrument-Control-GPIB-Serial/What-is-the-difference-between-doing-a-VISA-read-asyncronous-vs/m-p/60673?requireLogin=False -
RE: Reading measurement data from Keithley SourceMeter 2602A
As for speed of the source-measure sweep, pg 30 in the 2600B series datasheet gives an idea.
Max expected rate for source and measure operations into memory is 12000 per second.
The 2600A series will have similar capability.
In your pulse mode Python code, the pulse period would need changed to 83.3usec to attempt that 12K rate.
-
RE: Reading measurement data from Keithley SourceMeter 2602A
This is fairly common issue to coordinate the state of the Python code with the state of the instrument.
For the 200 point sweep, it clearly requires more time to conduct the sweep vs. the 100 point case.
The VISA session has a timeout parameter, in msec. Default value is 10 seconds for most VISA versions.
If your 200 point sweep needs more than 10 seconds to complete, then is very possible to receive a VISA timeout from the VISA read following your printbuffer() command.
There are a number of strategies to deal with this.
One would be simply to set the VISA session timeout to a larger number. Default is 10000 msec.
Downside of longer VISA timeout is that for real error conditions, it will take longer to raise the error.
Or introduce a time.sleep(seconds) on the PC side before asking for the data.
Alternately, it is possible to have the instrument assert an SRQ when the sweep completes and your Python can do status polling and ask for data only when the instrument signals that is has some ready for you.
Really depends on how fancy you want to get in the code. -
DC-DC Converter Efficiency test using two 2450 SourceMeter instruments
From old forum: https://forum.tek.com/viewtopic.php?f=263&t=143187
In general, the problem in the code was how the voltage sourcing SMU at input to the DC-DC converter was programmed.
Revised Python code for two 2450 attached.
-
RE: Problem Using DMM6500 with 2000-Scan Card
Looks like this one found a work around: https://forum.tek.com/viewtopic.php?f=617&t=142379 -
RE: DMM6500 Timer bug
In addition to checking your firmware level, also take note that when you edit the Timer settings, an Enable Timer button is displayed.
After the new delay value is entered, Press the Enable Timer button to apply those new settings to the timer.
The Enable Timer button then disappears until you again have changed one of the timer settings.
If you were operating from software control and using the timers, setting or editing timer values requires disabling the timer, set the various parameters and then enable it.
Seems the touch screen GUI is detecting a change is made and awaiting for you to apply them with the Enable Timer button. -
RE: AC Current Pulse with 2636 SMU
Over GPIB, a program can send abort command to halt an executing script.
For this older non-A and non-B model, the manual leaves me a little unclear on exact syntax.
Try: abort or abort() or smuX.abort()
For the A and B versions, the smuX.abort() is the proper syntax. -
RE: Keithley 590 C-V Analyzer: Applying the Calibration Capacitor Correction
Curious which test freq does your 590 have? 100KHz or 1MHz or both?
Which freq do you use?
For the calibration capacitor topic, do you have the 5907 standard caps? Or supplying your own caps?
Are you connected through the switch when doing the cable compensation with the standard caps?
You should be if you want the method to correct for it.
You mention an unwanted offset of up to 20pF when connected through a switch. Have you tried the zero method from section 3.10?
Your 590 is likely to be quite old. When was last time it was calibrated?
How are you determining the value of the reference caps?
In general, I’d try evaluating just the 590 with minimal cabling to the standard caps. If successful, then start adding in the other aspects such as the switch, etc.